Continuing the discussion from Building a bench supply from an ATX power supply? (A study in masochism):
My first step was to acquire a precision voltage reference and use it to standardize my DVMs.
(The following was copied from a message to @cyclicalobsessive)
====================
Calibrating my precision voltage reference and my DVMs.
The precision DC voltage reference:
-
Vcc = appx. 18v. (two 9v batteries in series)
-
Anything better than about 12v is OK. I vaguely remember that Vcc maxes out at either 40v or 60v for the Analog Devices programmable precision voltage reference IC
I paid less than $10 USD for it from China so I can’t complain. Maybe it’s not good enough for a HP precision Ceasum Atomic Time Standard, but it’s good enough for me.
It provides the following standard voltages based on jumper settings:
-
10.00
-
7.50
-
5.00
-
2.50
====================
First:
I went to Chip and Dip and measured the voltage reference with their Fluke DMM. They assured me that the meter was probably more accurate than my reference. Even if that’s not precisely true, I accepted as true that it was likely better than my el-cheapo DMMs.
The measured voltages using the Fluke DMM:
-
2.50 => 2.497
-
5.00 => 4.999
-
7.50 => 7.49 (change in precision)
-
10.00 => 9.99
Rounding to 2 decimal digits the difference was so small I decided to round to even voltages.
====================
Then, I measured the reference voltages with my two DMMs
-
A PMM-1000 DVM I bought somewhere.
-
A MT-009 Auto ranging DVM I bought from Bangood.
Measured voltages:
1. PMM-1000 DVM delta
* 2.50 => 2.46 - 0.04
* 5.00 => 4.94 - 0.06
* 7.50 => 7.42 - 0.08
* 10.00 => 9.90 - 0.10
Measured on the 20v scale
2. MT-009 Auto ranging DVM
* 2.50 => 2.51 + 0.01
* 5.00 => 5.02 + 0.02
* 7.50 => 7.51 + 0.01
* 10.00 => 10.02 + 0.02
Allowed to auto-range automatically
The conclusions:
-
The PMM-1000 has a linear - 0.02v offset per 2.5v. (zero appears to be right on the mark so it’s a gain offset at the high end.)
-
The MT-009 DVM appears to have a constant + 0.015v offset throughout the range of the voltage standard as noted by the last digit flipping back and forth.
With this data I can begin to calibrate the adjustable supplies on the bench power supply.
====================
Next steps:
-
Calibrate the variable voltage outputs on the bench supply.
-
Standardize the value of some high power resistances I bought.
-
Use these resistances to calibrate the current scale on the variable voltage outputs.